Calibrate load gauge
rblunt
Nov 2013
Canada
35
We have a gauge that is measuring the force we press with, whats the industry standard way to calibrate it? Our press gauge is capable of 11240 lbs.
If you have the load cell calibration sheet available and the calibration procedure for the signal conditioner and you work out the math and you can use shunt calibration method on your signal conditioner, that's the way I have and do perform calibration on all the load cells I have used.
with a traceable check gauge (just a simple big gauge that is calibrated.
or with a pressure calibration tool (it has a piston and you put weights on it, to getinto the range needed, if the weights are floating you can calculate the pressure from a table.
(i do have one without weights from DRUCK)
Another way I have used is a calibrated (by an outside firm) load cell that you put into the process and then load it up and compare the readings.
I will drive down toward Charleston to see 11000 lbs of NIST weights.
We have two steel pallets of 6 1000 pounders each, and then other breakdowns of 500, 250, 100, 50, 20, 10, 5 and 1 pounders. We need them to calibrate everything from large shipping scales to small ingredient weigh systems.
How else would you calibrate 5+ ton shipping scales?
The actual container for each group of 6 1000 pounder really isn't that big, the weights themselves are about a 14" cube with an attachment hook.
Here you go for some specifics:
http://www.scalesgalore.com/Rice_Lake_Class_F_Grip_Cast_Iron.cfm
I was told incorrectly, we have 4700lbs is the maximum according to ballscrew guys.
So the plan is to take a nitrogen spring, squeeze it with hydraulic press which has a gauge and then go push it with our actuator and equate the strain gauge reading to push applied.
Still not sure exactly if this is good enough because I'm relying on the hydraulic press gauge to be accurate.....
None of our stuff has to be certified for trade so that makes things simpler. If yours doesn't, then I wouldn't worry about NIST certified weights.
Going with that assumption, I recommend that you use what you have available, and check your results against your process.
Personally, you are better off with a reasonably simple procedure that checks your calibration that you can do on a regular schedule (to watch for drift), than a complicated procedure that only gets done when you suspect that there is a problem.
Hi guys, I am using LCSoft the calibrate load cell, the LCSoft send a specific value to PLC then i can use it and scale or calibrate in PLC Delta...
3,470
If any body could help me. 1) I have a load cell with 2mv/v output and 10-15vdc as power supply. How do I calibrate for 0-1000kgs.(for both...
14,062
8,596
E-OSULL
E
Hi all, Recently a client had an issue with their Hach sc100 and while I got the instrument back up and running, the analog outputs for the two...
1,570
Hi. Almost every time when working with PT100 probes the customers ask me "have the probes been calibrated?" The reason for the customers to ask...
18
5,083