I had a new one for me today. I got called to an X-Ray machine today because it wouldn't turn on. The X-Ray is complicated but the actual controls to turn it on are quite simple. A start - stop station run in series with customer interlocks (ours) and then to a contactor and two lights to indicate power on or power off.
Not having a drawing I started probing for voltages on the pushbuttons to find that the PB's, lights, and contactor all terminated on an unlabled PCB.
Looking at the PCB all I could find was a transformer, a couple of relays and some resistors. I was getting frustrated as to why those "(insert your own explicitive here) Europeans" would complicate something that should be simple with a mysterious circuit board.
Turns out the purpose of the board was to sense when one of the on or off light bulbs was burnt out and to prevent energizing the contactor if that was the case.
In as safety sensitive application like an X-Ray I think that is a great idea. I've just never seen it before. It solves the whole "what if the light bulb burns out" scenario.
Maybe I have been living in a hole the last 10 years. Is this common in the controls industry?
I am looking for either yes I see this all the time or no that's a new one on me responses.
Not having a drawing I started probing for voltages on the pushbuttons to find that the PB's, lights, and contactor all terminated on an unlabled PCB.
Looking at the PCB all I could find was a transformer, a couple of relays and some resistors. I was getting frustrated as to why those "(insert your own explicitive here) Europeans" would complicate something that should be simple with a mysterious circuit board.
Turns out the purpose of the board was to sense when one of the on or off light bulbs was burnt out and to prevent energizing the contactor if that was the case.
In as safety sensitive application like an X-Ray I think that is a great idea. I've just never seen it before. It solves the whole "what if the light bulb burns out" scenario.
Maybe I have been living in a hole the last 10 years. Is this common in the controls industry?
I am looking for either yes I see this all the time or no that's a new one on me responses.