Pressure transmitter cooling element/standoff
g.mccormick
Jul 2012
IN
961
I need to protect some pressure transmitters from hot (120°C) oil. Does anyone have a favorite transmitter cooling element/standoff they use for this?
Plain old impulse tubing is the answer. Tubing drops a lot heat because the tube is deadheaded with no flow through it.
The conventional rule-of-thumb myth is that impulse tubing drops 100-150 Deg (F or C?) per foot, but Kulite did a study and established a evidential rule of thumb: one foot of steel tubing, any diameter, will isolate a transducer from any temperature.
Kulite did the thermo math: https://kulite.com/assets/media/2018/01/AN8401HeatTube.pdf
I have witnessed this personally - 1 foot of inconel tubing through the wall of a furnace running at 1800°F and I could touch the compression fitting on the body of the transmitter and not get burned. There was a 3'x3' galvanized sheet metal shield between the furnace and the transmitter to prevent radiated heat from the furnace from overheating the transmitter, but combustion gas was cooled by that 1 foot of tubing.
Fill the tube with oil during assembly and locate the transmitter below the elevation of the tap, so that any air/gas rises and escapes from the impulse tube and does not accumulate in the impulse tube.
After getting some pricing and availability, I think pigtail siphons will work just fine. I'm only trying to protect from 120C oil.
Does anyone have experience wiring this type of analog pressure transmitter up? It's 4 wires. 3 blacks and 1 green which I assume is earth...
265
Hi All, Just wondering what the best/most accepted way to measure tank level off pressure transmitter. Pressure transmitter is located in line...
1,937
I'm trying to wire a Endress+Hauser PMP71 Pressure Transmitter to a Schneider Electric TM3AI2H Analog Input Module. The Pressure Transmitter is...
1,304
Hi folks, I am having a hell of a time getting an E+H PMP23 scaled correctly in a CompactLogix PLC. I have tried all sorts of data manipulation...
19
7,021
Hello Everyone , Hope you all are well and safe sound! I'm looking solution for the pressure transmitter which are installed in Blender or Lpg/...
2,885