Oakton Conductivity meter

Steve Bailey

Lifetime Supporting Member + Moderator
Join Date
Apr 2002
Location
The boondocks of Western Massachusetts USA
Posts
8,593
I'm trying to get the measured conductivity from an Oakton CON2700 conductivity meter into a PLC. The unit has an RS232 output for a printer. Instead of connecting to a printer I plan write a script in the machine's HMI to read the stream of ASCII characters coming out of the RS232 port, decode the conductivity value from that, and pass the value to the PLC.

I purchased a cable from Cole-Parmer that was listed in the Oakton manual, but as soon as I connect it to the serial port on my laptop the measured conductivity value jumps by a significant amount and becomes unstable.

Is there anyone familiar with the conductivity meter who can offer any suggestions?
 
Ground Loop.

It can be between the measured fluid, the analyzer or the RS-232 ground on the laptop.

If the measurement floats, like in a glass beaker, and the offset from the ground loop disappears, you're probably stuck, because the process is not at the same ground potential as the final data destination, the Laptop COM port.

Many times, the ground loop occurs when the laptop is charged by AC, but disappears when the laptop goes to battery power.

If the serial port is through a USB converter, there isolated USB/232 converters that will solve the problem, if that's where the ground loop is.

Industrial conductivity analyzers are probably isolated, have Modbus, panel mount and are not much more than the bench top lab analyzer, but maybe you need that lab unit.
 
I didn't know if one existed but I websearched for 'RS232 Isolator' and got lots of hits ranging from £4.00 on ebay to $200.00 from http://www.bb-elec.com. Can't advise which would be best for the job but as long as it is self powered and opto-isolated I think it would be OK.
 

Similar Topics

Hi All, We are looking for a new paperless recorder, our current ABB SM1000 is on its way out, the screen is difficult to see. There is a long...
Replies
13
Views
2,556
Hello, I have a Suntex EC-410 Conductivity controller that feeds a 4-20mA signal back to a compactlogix. I have the manual for this device but no...
Replies
4
Views
1,328
I've been asked to build a "simple" salinity monitor / controller for a chilled brine system. To simplify things, the owner prefers a tank-mounted...
Replies
4
Views
1,952
Hi Guys I need a bit of help here with my math! I have to measure the conductivity of caustic in a line. Now what I have been given by the...
Replies
15
Views
3,582
Hi: Replacing an old open board controller with a small PLC. I have a tank with 5 probes (1 reference and 4 different levels). I need an...
Replies
10
Views
5,795
Back
Top Bottom